shockrt
controllogix Interfacing with Blood Gas Analyzer (RAPIDPoint 500)
My client is a pharmaceutical company and they want to interface their Blood Gas Analyzer (which is a RAPIDPoint 500) to a Controllogix PLC via EtherNet. I have attached a Spec that I found, which gives the initial information you would need to do this. It says that the BGA can transmit data in ASCII. However, it doesn't list what bits I need to send to the device in order for it to send the PLC the data. I've tried contacting Siemen's (who makes the BGA) and they were useless. I also tried searching the RA knowledgebase.
Does anyone have any insight on how to communicate with a device in this way?
Thank you
Tyler
RP500_LIS_5_20_13_-_Rev_C_.pdf
I would get an ASCII to EtherNet/IP gateway to do this.    My favorites are the HMS Fieldbus "Communicator" (AB7072 or the new  HMS-EN2SE-R)and the Real-Time Automation 435NBX gateway.
The Communicator is probably faster, and could do more onboard parsing of the strings from the BGA instrument, but the 435NBX is dead-simple to configure and use.
This sounds like a really cool application.   Are they controlling a storage system that handles samples for the BGA ?
RP500 comm manual page 2-7:  Patient Sample Assay Transaction
- The analyzer notifies the host (LIS) that a patient sample analysis has begun or been cancelled.
- Host acknowledges
- analyzer notifies host that new patient data is available
- host acknowledges the msg
- host request the data and identifies the sequence number of the requested data
- analyzer sends data
- host acknowledges
